寻找 5 channel Peristaltic Pump (5 通道蠕动泵) 产家, 技术要求如下:
· The pump must be configurable to be either 3, 4 or 5 channel.
· The pump must have sufficient rollers to meet the requirements of the expansion/compression of the tube, but not less than 8.
· The rotor diameter must be between 50mm and 70mm diameter.
The range of tubes the Peristaltic pump must be able to operate with is as follows:
· Black/Black,
· Orange/Orange
· White/white,
· Red/Red
· Grey/Grey
· Yellow/Yellow
· Yellow/Blue
· Blue/Blue
· Green/Green
· Purple/Purple
· Purple/Black
Tube material to be used with the pump is as follows:
· STD PVC
· Marprene
· Viton
· Solvent Yellow (mod PVC)
Wall thickness of tubes to be compatible with the pump:
· 0.85-0.90mm
The pump design must be such that PVC Grey/grey tube running at 15rpm constantly should last 1 week.
· The pump presser arms must be optimized to operate with minimal pulsation on White/White & Grey/Grey tubes.
· Expansion of these tubes must be over 2 rollers.
· Compression of tubes must be gradual over 1 full roller pitch, without sharp transitions.
· Shape of expansion zone of the presser arms must be optimized to provide for constant volume change in the optimized tube over the 2 expansion rollers.
· Presser arms must be clamped independently, so that individual channels may be switched off.
· Clamps must be designed such that presser arms can be clamped & unclamped in a single handed operation
· When unclamped, the presser arm must fall away from the rotor.
· When unclamped, the gap between the presser arm and the rotor must be not less than 10mm at the pivot point, and not less than 30mm at the clamp end.
· Not withstanding the above 2 requirements, it must be possible to thread a tube around the drum past the presser arm closest to the drive motor, whilst the outer channels are all clamped.
· Presser arms must have a feature that ensures the tube is not able to migrate off the track of the arm.
· Clamping mechanism must have provision for operator adjustment.
· Tube holders must be provided which are a simple drop in type, capable of accepting all the tubes listed above.
· Tube holders must be designed such that the tube does not drop out of the holder if the tube stretches.
Normal operation will be 12-15rpm (approx 1.25ml/min on the breadboard). Pump must run at up to 50rpm.
· Pulsation: See presser arm requirements for pulsation requirements.
· The motor must be a 2 phase stepper motor.
· The motor must be capable of delivering sufficient torque to drive 5 channels of all tube material types, over the full temperature range and speed range.
5-45ºC from sea level to 3000m
5-35ºC from 3000-4500m.
· Initially volumes of 250 per annum are anticipated, growing to 500 after 2-3 years.